Extracted-ion chromatogram baseline
ID: extracted-ion-chromatogram-baseline
An extracted-ion chromatogram baseline is a model of the signal underlying a selected chromatographic peak in an extracted ion chromatogram. Subtracting it helps assign an extracted-ion chromatogram peak area to the feature of interest. The baseline need not be zero or constant, and it is not automatically the signal from a separate reagent blank.
Let be the signal from an extracted ion mass window and let be the baseline over . The assigned extracted-ion chromatogram peak area isThe time-ordered mass selection follows the IUPAC extracted ion profile definition; the baseline is an additional integration-model choice.[ref]
Extracted ion profile in IUPAC Gold Book. Entry 12410, signal from selected mass-to-charge values across time-dependent spectra.
An original example takes and a baseline rising linearly from 1 to 3 signal units. Its integrated area is signal-unit minutes. If the recorded trace has total area 20, the corrected area is 12. Using a constant baseline 1 would subtract only 4 and assign area 16: a 33% overestimate relative to 12. No recorded ion signal changed; only its allocation between baseline and peak changed.
If a fitted baseline is over a duration starting at , then the subtracted area is . Chromatographic baseline-area uncertainty therefore includes the joint uncertainty of the fitted intercept and slope:This is a direct application of correlated-input uncertainty propagation.[ref]
Candidate baseline regions should be inspected for neighboring peaks, broad tails and drifting background. Two overlapping compounds can generate one smooth trace; subtracting a convenient baseline cannot establish chemical specificity. Record the integration boundaries, baseline model and treatment of negative corrected values. Those choices are necessary to interpret or reproduce a reported extracted-ion chromatogram peak area.