Raman pump-on pump-off gain ratio

ID: raman-pump-on-pump-off-gain-ratio

A Raman pump-on pump-off gain ratio compares the probe spectrum with and without the narrowband Raman pump, after correcting detector background and normalizing by a reference. For positive dark-corrected probe values and reference values ,
Unity denotes no pump-induced change in this model; is the fractional gain. The primary FSRS experiment explicitly performs dark subtraction, reference normalization and the on/off division.[ref] These operations define a measured optical gain observable, not a direct count of Raman photons alone.
An original offset example has true probe readings with pump off and with pump on, with an unchanged reference ratio. The true gain is . A common unremoved detector offset of changes the raw ratio to
Reference normalization does not automatically remove an additive detector background. Uncertain subtraction also contributes uncertainty to the corrected ratio.
The shot-noise precision of a Raman gain ratio gives a separate statistical limit. If the four inputs are independent Poisson photoelectron counts with large means, first-order propagation gives
For weak gain and all four means approximately , . At , that is about : a one-percent gain is only about2.5times this standard uncertainty. These are chosen photoelectron-count assumptions, not an uncertainty specification for arbitrary digitized detector counts.
Shared laser fluctuations, read noise, uncertain references and non-Raman nonlinear effects require their own covariance and response model. The ratio can reveal a weak band only when the acquisition and background control justify its precision. The parent femtosecond stimulated Raman spectroscopy article explains the pump/probe apparatus and supplies a distinct reference-fluctuation example.