Recovery correction

ID: recovery-correction

A recovery correction adjusts an analytical result to account for an estimated loss of analyte during a specified preparation stage. If the recovered amount follows , with original amount and extraction recovery , the corrected estimate is . Its validity depends on whether the recovery estimate applies to that analyte, sample matrix and procedure.
For an original example, and give . The calculation estimates the amount before loss; it does not restore molecules to the sample. It also does not separately correct a mass spectrometry matrix effect unless that effect has been included consistently in the measurement model.
Recovery-correction uncertainty follows from the quotient, not just from the measured amount. First-order propagation of uncertainty gives
If the two inputs are independent, the covariance term vanishes. With stipulated standard uncertainties and , the two variance contributions are and , giving . The recovery estimate dominates this example; dividing by recovery has not created extra precision.
A recovery inferred from a spiked preparation need not reproduce the behavior of analyte originally embedded in a sample. Likewise, applying a single recovery at every concentration assumes a common model. Near-zero or very uncertain recovery can make quotient inference unreliable beyond the linear uncertainty approximation.
Reporting whether a correction was applied, its size and how it was derived makes results comparable. These reporting provisions appear in the harmonised recovery guidelines and their Codex adoption text.[ref]