Average power of a laser pulse train

ID: average-power-of-a-laser-pulse-train

The average power of a laser pulse train is its integrated radiant energy per observation time. If an interval contains complete pulses, with no uncounted edge contributions or continuous background,
Thus the relevant pulse repetition frequency is the actual emitted-pulse rate in the interval. Historical calorimetric calibrations inferred average power and pulse energy from total received energy, injection time and pulse rate.[ref]
An original counting example exposes a possible cancellation of errors. A nominal trigger produces 200 triggers in . Suppose only 180 complete optical pulses arrive, each carrying . Their total is , so the pulse contribution is , corresponding to an actual 90kHz pulse rate. Add a continuous optical background of , and a detector integrating everything reads .
Dividing that total by the nominal trigger rate would happen to give , despite both the missing pulses and the unremoved background. Removing the background but still using the nominal rate would instead infer , ten percent below the actual emitted-pulse energy. This missing-pulse bias in average-power inference cannot be diagnosed from the apparently reasonable quotient alone. A pulse-resolving monitor and a suitable background subtraction distinguish the two contributions.
For bursts or variable repetition, count the pulses in the same interval as the energy measurement. A slow integrating detector can establish their mean energy after these corrections but does not resolve pulse-to-pulse energy fluctuation or optical peak power. Laser pulse energy separates these observables.