Shared gauge-block reference covariance is the dependence between calibrated length estimates caused by using the same uncertain reference. It explains how a gauge-block reference uncertainty floor can remain in individual absolute lengths while canceling from their difference.
Let corrected estimates be and . Suppose the shared reference error has variance , the comparison errors have variances , and these three contributions are independent. The covariance matrix is
In particular, . The covariance follows from the shared input, rather than from whether the two comparison readings happened at the same instant.
For the contrast ,
Both reference sensitivities are exactly one in this additive measurement model, so their contributions cancel. General uncertainty propagation includes precisely this covariance term.[ref]
An original example uses , and . The individual standard uncertainties are about 22.36 and 25 nm, while the difference has standard uncertainty . Incorrectly treating the absolute length estimates as independent would give for the same difference.
For any linear combination , the reference contribution is . It vanishes for a difference but remains in an equal-weight average, where . Repeatedly averaging values carrying the same reference error does not produce independent reference information.
Unequal gauge-block reference sensitivities change the cancellation. If the reference enters as and , a difference retains reference variance . Shared temperature, comparator or wringing contributions also need their own coefficients and dependence assumptions. NIST's uncertainty procedure separates the master history from comparison-process variability.[ref]
Retain the calibration lineage and input relationships with the estimates. A pair of certificate uncertainty numbers alone does not establish their covariance or justify setting it to zero.