Write the magnetic flux linkage of the stationary pickup coil as , in SI units of weber. If every turn links the same magnetic flux , then ; the turn count is dimensionless. With a fixed winding and terminal convention, Faraday's law of induction gives the induced electromotive force
This is the fundamental theorem of calculus applied to magnetic flux linkage. Its units are , not volts alone. Reversing the leads reverses the sign. In a high-input-electrical resistance voltage measurement the input approximately follows this electromotive force; coil loading and the frequency response of the electronics require separate calibration.
Separate direct drive linkage from material linkage: . Even an empty pickup coil has . In an ideal long, uniformly coupled specimen with negligible demagnetizing field correction and eddy current shielding, a useful material approximation is
Here is specimen cross-sectional area, not the larger coil aperture; is vacuum permeability and the axial magnetization sampled by the coil. The units check directly: . Finite geometry and nonuniform magnetization generally require a calibrated coupling in place of . The exact linkage identity survives that failure of the uniform approximation.
As an original illustrative calculation, stipulate , , and rounded . Then
A uniform change over would give a rectangular pulse of ; over it gives with the same area. Thus pulse height mixes change size with change speed. These numbers describe an ideal model, not observed pulses or material constants.