RNA integrity describes the preservation of RNA molecules relative to the molecular lengths and structures required by a stated analysis. RNA degradation can leave fragments containing short target sequences even after long transcripts have been broken. Integrity is therefore distinct from total RNA amount, chemical purity, correct nucleotide sequence and preserved biological activity.
An electrophoresis trace measures a size-dependent distribution of material. For a total-RNA preparation, ribosomal RNA contributes prominent features, while smaller fragments can alter the trace as degradation proceeds. The RNA integrity number introduced by Andreas Schroeder and colleagues combines several trace features into an algorithmic score. Its 1-to-10 scale is not a percentage: a score of eight does not mean that of every transcript is intact. A single ribosomal-peak ratio also differs from the multi-feature score.[ref]
The RIN: an RNA integrity number for assigning integrity values to RNA measurements by Andreas Schroeder and others. Abstract, Feature selection, Discussion and Figure8
An original cleavage model explains why assay length matters. Suppose cuts form a Poisson process along an RNA strand, with mean density cuts per nucleotide. A required segment of length is unbroken when it contains no cut:At the stipulated density per nucleotide, a 100-nucleotide segment remains intact with probability , while a 1000-nucleotide region survives with probability . Detecting the short sequence consequently does not prove that the full long transcript survived. This is an illustrative independent-cut model, not a fit to a particular RNA sample or a conversion from an integrity score.
A motivating experiment is the 2008 comparison of RNA preservation during laser capture microdissection. Mathieu Clément-Ziza and colleagues measured RNA quality after staining and over the subsequent collection interval. Their tested argon-atmosphere and ethanol-based preparations stabilized the observed RNA quality for up to 90 minutes. Figure1 combines time comparisons with instrumental separation traces.[ref]
Stabilization of RNA during laser capture microdissection by performing experiments under argon atmosphere or using ethanol as a solvent in staining solutions by Mathieu Clément-Ziza and others. Abstract and Figure1: time-dependent measurements, sample scrapes and Bioanalyzer readout
An integrity assessment should state the specimen, handling history, measurement method and downstream requirement. Quantitative PCR directed at one short region and an analysis requiring an intact long transcript need not fail at the same degree of fragmentation. A stable total concentration cannot establish intactness, and a bulk quality score cannot certify every individual molecular species.