Simpson baseline integration for pulse-energy correction

ID: simpson-baseline-integration-for-pulse-energy-correction

Simpson baseline integration for pulse-energy correction subtracts a three-point estimate of integrated background from a measured gate integral. For a baseline on , Simpson's rule gives
It is exact for a quadratic or cubic baseline, provided the three values represent background rather than pulse-contaminated samples. For a four-times differentiable baseline, the remainder is for some .[ref]
For an original chosen quadratic background, take with beginning, midpoint and end values , and . Its integrated baseline is , whereas endpoint averaging would give . A raw gate integral of therefore gives after the three-point correction. Overestimating only the midpoint by would subtract an extra .
The uncertainty of a three-point baseline integral is separate from its quadrature error. With baseline-sample covariance matrix and , its variance is . Independent samples of common standard uncertainty give ; a perfectly shared additive baseline error of the same uncertainty gives instead. Covariance terms cannot be discarded.[ref]
A matched blank acquisition or independently measured background can supply the midpoint. Historical NIST pulse-energy apparatus used before/after baselines for background and thermal-drift compensation.[ref] The three-point example is an original extension, not a claim about that instrument's processing. Higher polynomial exactness does not compensate for an incorrectly identified baseline.