A bipolar gradient is a magnetic field gradient waveform with lobes of opposite sign. In nuclear magnetic resonance, such waveforms can control sensitivity to position and motion. Opposite signs alone do not guarantee equal signed areas, and an applied waveform must be distinguished from its effective encoding waveform when refocusing pulses change phase evolution.
For an original counterexample, apply for and then for . The gradient zeroth moment isThe waveform is bipolar but unbalanced.
If the two durations are instead equal to , and the effective lobes are contiguous, from time 0 to and from to , then but the gradient first moment is . Thus cancellation of stationary-position phase can coexist with sensitivity to constant velocity. The sign follows this stated ordering.
Bipolar pulses also appear outside flow imaging. A 1989 stimulated echo study addressed background gradients that distort diffusion measurements. Alternating pulses could cancel a cross term under specified conditions; tests used proton signals from homogeneous and granular samples.[ref]
Pulsed field gradient stimulated echo methods for improved NMR diffusion measurements in heterogeneous systems by R. M. Cotts and others. Primary abstract: background-gradient reduction, conditional cross-term cancellation and measurements
This is a sequence-specific result, not a promise that every bipolar pair removes every imaging error. New to topics? Read the docs here!