A reference material is established as suitable for a specified measurement use because the relevant properties are sufficiently homogeneous and stable. The requirement concerns particular properties and an intended use, rather than every possible attribute of the material.[ref]
International vocabulary of metrology by JCGM. VIM3entry5.13, definition
Reference material homogeneity concerns variation within or between the portions used; reference material stability concerns maintaining the relevant properties over time and under stated conditions.Reference materials can support measurement precision control even without an assigned quantity value. An assigned value is needed when the material serves in measurement calibration or a check of measurement trueness.[ref]
International vocabulary of metrology by JCGM. Entry5.13, note2
A certified reference material adds authoritative documentation specifying property values with associated uncertainties and traceabilities obtained through valid procedures.[ref]A concrete physical example is the archived 2009 certificate for National Institute of Standards and Technology SRM 1922, a mineral oil characterized for refractive index. Its characterization used a precision spectrometer and goniometer arrangement with the minimum deviation method, oil in a hollow optical prism, and water-bath temperature control. Measurements covered several wavelengths and temperatures; multiple portions were examined to assess homogeneity.[ref]
Certificate for Standard Reference Material 1922 by National Institute of Standards and Technology. Certificate issued 24 June 2009, p1, Certified Values and Uncertainties
The certificate's Table 1 reports, at wavelength 589.3 nm and temperature , refractive index 1.46945 with expanded uncertainty and coverage factor 2. This wavelength's value was obtained through a dispersion fit. The listed temperature coefficient is .[ref]
Certificate for Standard Reference Material 1922 by National Institute of Standards and Technology. p1, fitted 589.3nm values and coverage factor; p2, Table1
An original calculation using that linear coefficient predictsThe predicted change, in magnitude, is larger than the quoted expanded uncertainty at the reference conditions. Temperature therefore belongs with the assigned optical value; it is not an incidental label. This calculation illustrates the certificate's model, rather than supplying a new certified value or uncertainty.
Record the material identity, certificate version, property, wavelength, temperature and handling history with a measurement result. The archived certificate supplies a documented example of those links, rather than a claim about a currently available product or the validity of a particular bottle today.