Resistance thermometer self-heating

ID: resistance-thermometer-self-heating

Measuring electric current creates Joule heating, raising the sensing temperature. At fixed bath temperature, compare settled readings at and and extrapolate the approximately linear dependence toward zero current.[ref]
Under the local model ,
For synthetic readings at and at , the extrapolated value is . The lower-current excess corresponds to in the Pt100 local linear model. At , dissipates ; power alone does not specify a temperature rise without a thermal coupling model. These stipulated values are not measurements.
Repeat the comparison after settling and test the approximation if necessary. Changing from liquid to air changes heat transfer; a self-heating correction need not transfer unchanged. The saturation-pressure equilibrium measurement provides an existing example of controlling this disturbance in resistance thermometry.