Noise cost of polynomial baseline cancellation

ID: noise-cost-of-polynomial-baseline-cancellation

The noise cost of polynomial baseline cancellation is the change in random measurement uncertainty when extra background trends are removed. For a linear pulse estimate and reading-error covariance matrix ,
The moment conditions for baseline-drift cancellation constrain to reject specified polynomial backgrounds. They do not make the remaining random error vanish.[ref]
For an original comparison, assume a pulse occupies only the central sample and the other samples contain background alone. Three samples at with weights cancel constant and linear backgrounds. Five samples at can instead use
These weights preserve the central signal and have zero moments of orders . For independent reading errors of equal variance , the three-sample variance is , while the five-sample variance is
With and , the standard uncertainties are approximately and . The second design removes quadratic and cubic drift as well, but its larger reference weights increase noise. This compares two specified estimators; it is not a claim that collecting more background data must worsen precision.
NIST's pulse-energy apparatus used before/after baseline measurements for background and thermal-drift correction.[ref] The polynomial comparison above is an original model for choosing such measurements. Choosing minimum-noise weights under baseline moment constraints can exploit additional reference samples and their measured correlations. Pulse leakage into a reference or drift outside the cancelled polynomial family still creates bias.