Spirometer calibration verification

ID: spirometer-calibration-verification

Spirometer calibration verification checks whether a spirometer measures a known input within specified limits. It differs from measurement calibration, which establishes the relation between input and indication, and from an adjustment that changes that relation. A satisfactory verification applies to the tested equipment, assembly and conditions; it does not validate a person's expiration maneuver.
The primary 2019 technical statement specifies daily verification with a 3-L syringe cycled over multiple flows. Its stated measured-volume accuracy requirement is ±3%, and a filter used for testing is also used during verification.[ref] The known volume is useful precisely because it can be delivered at different speeds.
An original gain-and-offset model explains this design. If measured flow is , integrating one transfer of known volume over duration gives
For the same at two durations,
A single transfer cannot distinguish gain from zero offset in this model; two different durations can.
Take stipulated 3.000 L transfers lasting 1 s and 6 s, reported as 3.020 and 3.120 L. The inferred offset is , while . The slow transfer is 4% high, exceeding the cited 3% limit. If both transfers instead read 3.060 L, this model gives and : a 2% gain error. These are constructed instrument responses, not a calibration record for an actual device.
Identifying flow-sensor gain and offset with timed transfers relies on the model assumptions. Nonlinear response, leaks or an inaccurately known syringe volume can produce other patterns. Reference-volume uncertainty belongs in the uncertainty budget, and repeated transfers share it rather than independently re-estimating it. A test at several flows is therefore more informative than repeating one convenient delivery speed.

New to topics? Read the docs here!