Manometer height-to-pressure calibration

ID: manometer-height-to-pressure-calibration

The ideal measurement calibration is a straight line through the equal-pressure zero. For the stipulated water mass density and gravitational acceleration above, its slope is per vertical millimetre. A lower liquid mass density gives a larger length change for the same pressure, improving the height-reading contribution to measurement resolution but requiring more vertical space for a given pressure range. The pressure scale depends on the actual mass density, which depends on temperature; here is a chosen model value, not an exact property of all water samples.
Figure 1.
Calculated water-column pressure scale and a two-meniscus rounding model
. The left panel is the theoretical measurement calibration for and . The right panel models two liquid meniscus heights read separately to the nearest millimetre: each rounding measurement error is assumed independently uniform between and . Their difference gives a triangular probability density function for the inferred pressure error. The bounds and standard uncertainty describe this constructed model, not observed instrument data. Details.
Suppose each liquid meniscus height is rounded to the nearest scale division . Each height measurement error lies within , so the error in their difference lies within . The resulting pressure error is bounded by . To quantify a standard uncertainty, make the additional model assumption that the two rounding errors are independent random variables, each having a continuous uniform distribution on . Each then has variance ; adding the two variances gives
Here denotes standard uncertainty. The worst-case bound and standard uncertainty answer different questions. At they are respectively and of the reading. Shared reading bias or a different reading method can violate the assumed independence.
If the mass density, gravitational acceleration and height difference also have small independent standard uncertainties, a first-order estimate follows from the partial derivatives of :
This displays why a finer scale does not remove uncertainty in mass density or gravitational acceleration. A pressure transducer compared against the column needs its own differential-pressure calibration to establish an electrical response.