Detector linearity test
ID: detector-linearity-test
A detector linearity test checks whether changes in a detector's output remain proportional to changes in its incident signal over a stated operating range. For a photodetector with a fixed wavelength, geometry and readout gain, a linear response model iswhere is radiant power, is the baseline and is the appropriate response coefficient. A detector baseline subtraction removes the offset before comparing with . Linearity concerns the shape of that relation; an incorrectly calibrated constant can still give a linear detector.
A documented apparatus uses a beam splitter to illuminate two photodiodes at approximately a ten-to-one intensity ratio. NIST's measurement at compared responsivity with its low-power value, with the aperture filled and uniformly irradiated. The report explicitly notes dependence on irradiation geometry.[ref]
Spectroradiometric Detector Measurements: Ultraviolet, Visible, and Near-Infrared Detectors for Spectral Power by Thomas C. Larason and others. §9.1, printed pp. 85–86 and Figure 9.2
A detector's response to a small concentrated spot need not reproduce its response to the same total power spread over the aperture.For an original numerical example, stipulate baseline-corrected electric currents of , and at input powers , and . Normalizing the apparent spectral responsivity to the first point givesThe last point has three percent less response per unit power than the first. A straight line through the origin with slope would predict there instead of .
A useful test varies incident power while monitoring source stability and holding the beam position, illuminated area and electronic settings fixed. Record blocked-beam readings as well as illuminated readings. A residual that exceeds the combined measurement uncertainty identifies a departure from the tested model; it does not by itself locate the cause in the photodiode rather than an amplifier or other readout stage.
For pulses, average power alone may hide a nonlinear response to high peak power. A continuous-illumination test therefore does not automatically validate pulse energy measurement with a photodiode at a different pulse duration or spot size. Test the illumination and electrical range actually used, and distinguish a change in slope from a change in baseline.