Luis Chioye writes about precision analog-to-digital converters and their measurement circuits. The author biography in a 2013 Texas Instruments reference design identifies him as an applications engineer in its Precision Analog group, with degrees in electrical engineering from the University of Arizona and Walden University/NTU School of Engineering and Applied Science.[ref]
Precision Thermocouple Measurement with the ADS1118 by Mike Beckman and Luis Chioye. §9, p. 23, author biography at the time of publication
His work includes ratiometric measurement with resistance thermometers: sharing an excitation electric current can cancel its variation, but the input and reference filters must preserve that shared variation consistently.[ref]
RTD Ratiometric Measurements and Filtering Using the ADS1148 and ADS1248 Family of Devices by Luis Chioye. Abstract and filter-design discussion, matched signal and reference paths
This is the engineering distinction captured by ratiometric cancellation with matched signal and reference filters.With Mike Beckman, Chioye also published a thermocouple reference design. Its reference-junction temperature sweep in thermocouple testing keeps the measurement junction in a stable bath while changing the temperature of the connector and converter board; the acquisition computer stays outside the temperature-controlled chamber. This isolates reference-side effects from changing the temperature being measured.[ref]
Precision Thermocouple Measurement with the ADS1118 by Mike Beckman and Luis Chioye. §7, p. 20, Figure 17
Reference-junction temperature sweep: keep the measurement bath fixed while varying the connector and converter temperature. Original schematic of the test principle.
Details. An original ideal calculation explains the test. Let be the thermocouple electromotive force response from the Seebeck effect relative to . With isothermal connector junctions at , the measured electric potential difference is . Cold-junction compensation reconstructs before applying the inverse function of the response. In the stipulated linear model with , a fixed gives at and at . Correct compensation gives the same in both cases. These are constructed values, not measured performance or a calibration table for a particular thermocouple.
Cold-junction thermal coupling matters because the temperature sensor must represent the connector temperature. A precise converter reading cannot by itself remove an unmeasured temperature difference between those locations.[ref]
Precision Thermocouple Measurement with the ADS1118 by Mike Beckman and Luis Chioye. §2, p. 5, isothermal connector junctions and their temperature measurement