Platinum resistance thermometer lead resistance error
ID: platinum-resistance-thermometer-lead-resistance-error
With two shared leads of electrical resistances , Ohm's law givesUsing the Pt100 local linear model, positive lead electrical resistance appears as a warmer temperature. At true , stipulated leads of each change into , producing : a calculated bias of about . Subtracting a measured loop correction works only while that loop electrical resistance remains appropriate; cable temperature changes can invalidate it.
Four-terminal sensing separates the electric current and electric potential difference leads. Negligible sensing electric current suppresses sensing-lead drops; force-lead drops lie outside the interval. Real voltmeter loading, connections, and leakage still need checking.[ref]
For a stable offset from the thermoelectric effect, two signed electric potential differences obey and . The current-reversal calculation in four-terminal sensing yieldsAt stipulated , and , and recover exactly in this model. If the offset or sample temperature changes between readings, cancellation is incomplete. Reversal also leaves Joule heating unchanged because it depends on .